Question:
i would like to know about thermo couple?
kishore_sai_kishore
2006-05-27 22:45:32 UTC
it is like the couple must be of high resistance & it must have high boiling points however i.e.,aroundegree centigrade
Five answers:
kishore r
2006-05-27 22:56:45 UTC
In electronics, thermocouples are a widely used type of temperature sensor and can also be used as a means to convert thermal potential difference into electric potential difference. They are cheap, interchangeable, have standard connectors, and can measure a wide range of temperatures. The main limitation is accuracy; system errors of less than 1 °C can be difficult to achieve.



A thermopile is a group of thermocouples connected in series.



Principle of operation

In 1821, an Estonian physicist named Thomas Johann Seebeck discovered that when any conductor (such as a metal) is subjected to a thermal gradient, it will generate a voltage. Any attempt to measure this voltage necessarily involves connecting another conductor to the "hot" end. This additional conductor will then also experience the temperature gradient, and develop a voltage of its own which will oppose the original. Fortunately, the magnitude of the effect depends on the metal in use. Using a dissimilar metal to complete the circuit will have a different voltage generated, leaving a small difference voltage available for us to measure, which increases with temperature. This difference can typically be between 1 to about 70 microvolts per degree Celsius for the modern range of available metal combinations. Certain combinations have become popular as industry standards, driven by cost, availability, convenience, melting point and chemical properties, stability, and output.



It is important to note that thermocouples measure the temperature difference between two points, not absolute temperature.







In most applications, one of the junctions — the "cold junction" — is maintained at a known (reference) temperature, while the other end is attached to a probe. For example, in the image below, the cold junction will be at copper tracks on the circuit board. Another temperature sensor will measure the temperature at this point, so that the temperature at the probe tip can be calculated.



Having available a known temperature cold junction, while useful for laboratory calibrations, is simply not convenient for most directly connected indicating and control instruments. They incorporate into their circuits an artificial cold junction using the "other" thermally sensitive device (such as a thermister or diode) to measure the temperature of the input connections at the instrument, with special care being taken to minimize any temperature gradient between terminals. Hence, the voltage from a known cold junction can be simulated, and the appropriate correction applied. This is known as cold junction compensation.



Usually the thermocouple is attached to the indicating device by a special wire known as the "compensating" or "extension" cable. The terms are specific. "Extension cable" uses wires of nominally the same conductors as used at the thermocouple itself. These cables are less costly than thermocouple wire, although not cheap, and are usually produced in a convenient form for carrying over long distances - typically as flexible insulated wiring or multicore cables. They are recommended for best accuracy.



"Compensating cables" on the other hand, are less precise, but cheaper. They use quite different, relatively low cost alloy conductor materials whose net thermoelectric coefficients are similar to those of the thermocouple in question, but which do not match them quite as faithfully as extension cables. The combination develops similar outputs to those of the themocouple, but the operating temperature range is restricted to keep the mis-match errors acceptably small.



The extension cable or compensating cable must be selected to match the thermocouple. It generates a voltage proportional to the difference between the hot junction and cold junction, and is connected in the correct polarity so that the additional voltage is added to the thermocouple voltage, compensating for the temperature difference between the hot and cold junctions.



Ref: "Guide to Thermocouple and Resistance Themometry" pp20 Iss 6.0 TC Ltd.









Voltage-Temperature Relationship

The relationship between the temperature difference and the output voltage of a thermocouple is nonlinear and is given by a polynomial interpolation.







The an values are given to n between 5 and 9.



To achieve accurate measurements the equation is usually implemented in a digital controller or stored in a lookup table. Some older devices use analog filters.





Different types

A variety of thermocouples are available, suitable for different measuring applications (industrial, scientific, food temperature, medical research, etc.).



Type K (Chromyl (Ni-Cr alloy) / Alumel (Ni-Al alloy))

The "general purpose" thermocouple. It is low cost and, owing to its popularity, it is available in a wide variety of probes. They are available in the −200 °C to +1200 °C range. Sensitivity is approximately 41 µV/°C.

Type E (Chromel / Constantan (Cu-Ni alloy))

Type E has a high output (68 µV/°C) which makes it well suited to low temperature (cryogenic) use. Another property is that it is non-magnetic.

Type J (Iron / Constantan)

Limited range (−40 to +750 °C) makes type J less popular than type K. The main application is with old equipment that cannot accept "modern" thermocouples. J types cannot be used above 760 °C as an abrupt magnetic transformation causes permanent decalibration. Type J's have a sensitivity of ~52 µV/°C

Type N (Nicrosil (Ni-Cr-Si alloy) / Nisil (Ni-Si alloy))

High stability and resistance to high temperature oxidation makes type N suitable for high temperature measurements without the cost of platinum (B, R, S) types. Designed to be an "improved" type K, it is becoming more popular.

Thermocouple types B, R, and S are all noble metal thermocouples and exhibit similar characteristics. They are the most stable of all thermocouples, but due to their low sensitivity (approximately 10 µV/°C) they are usually only used for high temperature measurement (>300 °C).



Type B (Platinum-Rhodium/Pt-Rh)

Suited for high temperature measurements up to 1800 °C. Unusually type B thermocouples (due to the shape of their temperature-voltage curve) give the same output at 0 °C and 42 °C. This makes them useless below 50 °C.

Type R (Platinum /Platinum with 7% Rhodium)

Suited for high temperature measurements up to 1600 °C. Low sensitivity (10 µV/°C) and high cost makes them unsuitable for general purpose use.

Type S (Platinum /Platinum with 10% Rhodium)

Suited for high temperature measurements up to 1600 °C. Low sensitivity (10 µV/°C) and high cost makes them unsuitable for general purpose use. Due to its high stability type S is used as the standard of calibration for the melting point of gold (1064.43 °C).

Type T (Copper / Constantan)

Suited for measurements in the −200 to 350 °C range. The positive conductor is made of copper, and the negative conductor is made of constantan. Often used as a differential measurement since only copper wire touches the probes. Type T thermocouples have a sensitivity of ~43 µV/°C



Thermocouples are usually selected to ensure that the measuring equipment does not limit the range of temperatures that can be measured. Note that thermocouples with low sensitivity (B, R, and S) have a correspondingly lower resolution.



Applications



Heating appliance safety

Many gas-fed heating appliances like ovens and water heaters make use of a pilot light to ignite the main gas burner as required. If the pilot light becomes extinguished for any reason, there is the potential for uncombusted gas to be released into the surrounding area, thereby creating both risk of fire and a health hazard. To prevent such a danger, some appliances use a thermocouple to sense when the pilot light is burning. Specifically, the tip of a themocouple is placed in the pilot flame. This thermocouple electrically operates the gas supply valve responsible for feeding the pilot. So long as the pilot flame remains lit, the thermocouple remains hot and holds the pilot gas valve open. If the pilot light goes out, the temperature will fall along with a corresponding drop in electricity, removing power from the valve. Unpowered, the valve will then automatically shut off the gas, halting this unsafe condition.



Many systems (Millivolt control systems) extend this concept to the main gas valve as well. Not only does the electricity created by the pilot thermocouple activate the pilot gas valve, it is also routed through a thermostat to power the main gas valve as well. Such a system requires no external source of electricity for its operation and so can operate during a power failure, provided all the related system components allow for this. Note that this excludes common forced air furnaces because external power is required to operate the blower motor, but this is feature is especially useful for unpowered convection heaters.



A similar gas shut-off safety mechanism using a thermocouple is sometimes employed to ensure that the main burner ignites within a certain time period, shutting off the main burner gas supply valve should that not happen.



Out of concern for wasted energy, many newer appliances have switched to an electronically controlled pilot-less ignition, also called intermittent ignition. This eliminates the need for a standing pilot flame but loses the benefit of any operation without a continuous source of electricity.



Radioisotope thermoelectric generators (RTGs)

Thermocouples can also be applied to generate electricity in radioisotope thermoelectric generators.





Sleep Medicine

Thermocouples are one type of airflow sensor used in a polysomnography.



Steel Industry

Type B,S,R and K thermocouples are used extensively in the steel and iron industry to monitor temperatures and chemistry throughout the steel making process.
?
2016-05-20 10:47:00 UTC
It is based on the fact that when you heat a metal (wire in this case) electrons flow with respect to the heat source - but how do you measure it? If you weld two dissimilar metals together and heat the junction, the difference in the behavior of the electrons in different metals on heating results in a small but measurable EMF that increases with temperature. The actual EMF to be measured depends on the metals and on the difference in temperature between a cold point and a hot point. [If the measuring point were as hot as the hot junction, there would be a countervailing voltage developed to the middle of the wire.] We can make the cold point (junction) temperature fixed with an ice bath or some other manner, but in modern instruments we just measure the temperature at the point we are measuring the voltage. Because of resistance in the wires and so forth, instrumentation is made to have as little current as possible [high impedence] so just voltage is sensed. The heat pushes on the electrons, effectively. There is a kind of flame detector that uses one wire heated at the end by the flame and the electrical connection to ground is made through the flame plasma to the frame of the burner. Not sure what you are doing with the potentiometer in the thermocouple - the voltages are quite low and you need a strong heat source and a sensitive meter.
p
2006-05-27 23:11:31 UTC
thermo couple is a couple of two different metals



that is used for generation of electricity by causing a temperature difference between the two junctions



i don't think there are restrictions of resistance and boiling points



to choose a thermocouple however some series of metals is taken into account
Gary H
2006-05-27 23:02:22 UTC
a thermo couple is a safety device that keeps a main gas burner like on a water heater from turning on unless the pilot light is burning to ignite the main burner. the pilot burns against a metal rod that puts out a small electrical charge that goes to the main burner gas release. if it is off the burner will not let gas be turned on. if pilot is on,sends signal ok to start. sometimes they go bad and the pilot is burning but wont come on. solution, just replace it .keeps your house from filling up with nat gas without burning
faaajsh
2006-05-27 23:37:41 UTC
a thermo couple is a device that basic factor of that is using two type of me tall coupled together and their lengths increases in various point temperature and thus when one of them increases its length forced another them and it can causes a change in shape of them and we can use from it for on or off of a system